I am trying out the new version of local refinement implemented in 3.1 - very excited to see if it improves!
I launched two identical runs, one with an initial lowpass of 8 Å, and the other with an initial lowpass of 40. Both worked using the previous version of local refinement, but in the new version, the one with a 40 Å initial lowpass failed with the attached error (the other one is still running).
Actually I think there is something wrong with both jobs. In both cases, the “Alignment map A/B” entries are empty, as is the second half of the noise model initialization (the first step looks okay). The FSC filtered half maps are also empty.
Any idea what could be wrong here? All the inputs were used for a “normal” local refinement job without issue.
Thanks for reporting, and apologies that you’re encountering this error - we haven’t yet seen this in testing! Something does look like its going wrong with generating the initial alignment maps - the plot shows the alignment maps as empty and the rest of the job will definitely fail if that’s the case. Do the “Input Mask Real Space Slices” and “Initial real/fourier space slices” plots like as you’d expect, or are they all empty too?
If they look fine, the error could be related to the input mask – new local refinement has static masking on by default, so it doesn’t generates a new mask by default, whereas the old version would always do this (until we fixed a bug that ignored the “static” masking parameter in v3.0). Could you try re-running the new local refinement jobs, while setting the masking parameter to “dynamic” and setting the “Dynamic mask start resolution (A)” parameter to anything large (like at least 40Å, if you’re starting from a 40Å initial resolution).
That was indeed the issue. Setting masking to dynamic, it now runs correctly. There was nothing particularly exotic about the input mask though, so I am a little puzzled as to what was causing the problem.
Perhaps this is it - the mask I was using was a binary mask, generated with relion_mask_create and imported, because in the previous version I knew it was using dynamic masking. Perhaps if a hard edged binary mask is provided as a static mask, that causes the problem?
Hi @olibclarke,
.
Ah yes – you’re right, static masking always expects a soft mask and I believe this has to be in floating point format, rather than unsigned int or otherwise. If you want to use static masking, what you can do is take the input binary mask that previously would give an error, pass it through a Volume Tools job, and re-do the process of thresholding/padding (e.g. you could just set the threshold to 1 and set the dilate parameter to 10 for a 10 pixel soft edge). This should re-generate a new mask in the same way that dynamic masking does, and then that mask should work with the new local refinement job. Glad that it is working now, please let us know if you run into any other errors!
Thanks @mmclean, it is running, but it seems to be extremely slow - my jobs have been sitting at the “local cross validation” step for the last 8hrs. Granted this is a large (512px) volume, but I don’t think it should take this long… any suggestions to diagnose? Output of htop and nvida-smi attached.
Hi @olibclarke have you run non-uniform refinement (new) on same/similar data? Did you see any issue there? It’s the same code that is running in local refinement now
The FSC seems to go bananas after the first iteration (at least with a 40Å initial lowpass). This is with the new local refinement, with NU-refine and marginalization on.
Is it possible this is caused by using a binary mask to initialize dynamic masking? Perhaps it is not using the correct mask? Has anything like this been seen during testing?
Hmm, certainly strange. How many particles did you use, and how big would you estimate the masked area to be? Did the FSCs for the run with initial lowpass of 8 Å look any different? Also, is this with dynamic masking on? With using dynamic masking and starting from a very low resolution (40 Å), the mask can change quite a lot from iteration to iteration which is generally undesirable.
In our testing, we’ve found results with static masking to work better, at least in terms of avoiding overfitting, which is why the default mask parameter was set to “static” —perhaps trying the job with a static mask could alleviate some of these issues? As well, we’ve also seen better results starting from resolutions in the range 8 - 20 or so, so we decided to increase the default to 12 Å
Edit: Yeah, perhaps the dynamic mask could cause issues here, especially if starting from such a low resolution (then, inside the masked area, the density might vary slowly over space, meaning there’s basically nothing to align to except the mask itself – this is part of reason why for the other refinements, the dynamic mask start resolution is set as 12 Å by default).
Edit 2: Actually yes – I think the error is with initializing the dynamic mask with a binary mask. The initial mask is used for the tight mask FSC computation and so the tight FSC curve is very high because the volume is being sharply cut off, meaning basically all of the high res info is just going to be ringing in fourier space from the density being clipped by the binary mask (which is the same for both half-maps, explaining the high correlation). For best results, either static masking should be used, or dynamic masking, starting from a relatively high res of 12 Å (and in either case, soft masks are recommended).
It looks like this by iteration 1 whether the lowpass is 40 or 8 - I have used the same dataset with the original local refinement procedure and have not seen anything like this using the same mask and same parameters. The dataset contains 360k particles.
I am trying now with a static mask. Yes, I generally use initial lowpass resolutions in that range also.
Replacing with a static (soft) mask does the trick, thanks! I would note though that this binary mask worked just fine in the previous version of local refinement, starting from either initial lowpass value, whereas in the new version it works with neither - I think something must have changed in the dynamic masking parameters. I still see this behaviour when starting with an initial lowpass of 8 Å, so it is not all derived from filtering too aggressively to begin with.