I am trying to use the new Micrograph Denoiser and see these circles in the middle of my micrograph (with default setting). I am puzzled what are these, they are not present in my original micrograph after patch motion correction. I was wondering if I can change some parameters in input settings to fix these.
@Anamika1 If you push the Show from top button under Event Log, you should be able to scroll down to a section with learning progress plot and two (split) panels of micrographs like
Hi, Please see below the example of motion correction file for this data and also I did not use the Detector defect file. By the way this data was collected on falconIV detector.
I believe this problem results from a hot pixel. When we apply the CTF to the micrograph before denoising it, a hot pixel will essentially recapitulate the CTF, resulting in the rings you can see there. I believe the easiest way to fix this is to reach out to your collection facility with this information and ask them for a defect file.
This hot pixel also seems to be affecting your motion correction results (note that the stationary patches are in the same place as the rings). I would therefore recommend getting a defect file even if you decide not to move forward with denoising.
I wish I had an easier fix for you, but I hope that is helpful!